Many thanks to all of you who responded to my call for help in dealing
with the 16-Hz noise I was encountering in my sound booth. I was
surprised to see how many people have had similar problems, because it's
something I've never run into before (though I guess I've only
calibrated in a total of 6 booths over my career). 

I tried several of the suggested solutions, including changing the
orientation of my coupler and placing the coupler on a soft surface
(neither of those helped). I checked the spectra of my speech and babble
calibration signals and determined that C weighting would have no impact
on them. Unfortunately, the noise level was still close to 70 dB C. 

Finally, skimming Arlinger (1997) made me think to try my **2-cc**
coupler - and that solved the problem. Measured in this coupler, the
noise is 40 dB C or 50 dB SPL, low enough for me to measure my signals
at my actual test levels. 

All of this raises a larger question about calibration, which I'll pose
in a new thread. 

Cheers